Form Wash L flammable solvent FAQs
The Form Wash L complies with a variety of safety and quality standards. This article is designed to answer any questions related to using flammable solvents with the Form Wash L and to address any concerns with the machine’s risk of explosion.
Note:
For the complete safety instructions, please visit Manuals and documentation article to download the Form Wash L manual.
How does Form Wash L address the ignition risk of solvent vapors?
Many hydrocarbon solvents, including those used to post-process SLA prints, are flammable. The Form Wash L is designed to mitigate the risk of the ignition of solvent vapors within the unit with third-party verification of compliance to UL/IEC61010-1 standards relevant to risk of ignition and explosion. The Form Wash L bears an ETL Listed Mark and is CB certified for UL/IEC61010-1, which provides standards for built-in features to reduce the risk of ignition and to mitigate explotion severity and effects.
Is the Form Wash L fire safe?
The Form Wash L abides by regulatory standards, but it is critical to properly assess the surrounding work area for safety risks. Use components and equipment listed by nationally recognized testing labs to ensure safety from fire or explosion hazards. For equipment recommendations, see the Formlabs-approved distributors for supplies (SLA) article.
The flash point of a solvent is the lowest possible temperature at which it forms enough vapor such that it can ignite if exposed to an ignition source. Isopropyl alcohol (IPA) has a low flash point of 53 °F (12 °C), so take special care to ensure that ignition of vapors does not occur.
Formlabs recommends using flammable solvents such as IPA to clean your SLA prints of excess resin. Assess your operating area to ensure that there are no nearby sources which can ignite the IPA fumes. Ignition sources include, but are not limited to, open flames, hot surfaces, and electrical sparks. If working in an enclosed area, ensure that the area is well-ventilated. Take care when filling the wash bucket with solvent, when both lids are open, and when printed parts are drying. These are the times when vapor concentrations are highest.
Always abide by your company's safety protocols and local regulations to ensure that flammable solvents are handled safely.
What is the recommended solvent for the Form Wash L?
Working with flammable solvents like IPA poses a safety risk to personnel and property. Understand the risks and mitigate them with appropriate safety guidelines.
If your company does not have guidelines for safely handling flammable solvents like isopropyl alcohol (IPA), Formlabs recommends using a non-flammable solvent such as tripropylene glycol monomethyl ether (TPM) to wash SLA prints with the Form Wash L.
If using a flammable solvent such as IPA, take proper precautions to reduce the risk of fire or explosion.
Formlabs does not manufacture solvents. Always consult the safety data sheet (SDS) from your solvent supplier as the primary source of information for storage and handling.
What features of the Form Wash L are designed to mitigate the risk of explosion inside the unit?
The Form Wash L features used to mitigate isopropyl alcohol (IPA) flammability include the following:
  • Double lids to ensure vapors are contained by at least one physical barrier
  • All-metal tank construction
  • No use of ultrasonic agitation
  • Build platform tilts while drying to retain more IPA in its liquid state and less as vapor
  • Motion provided by stepper motors to eliminate sparks associated with DC motors
Are there specific features designed to mitigate the risk of explosion within the wash bucket?
The solvent monitor was evaluated for intrinsic safety and is rated for:
  • Class I, Division 1 and/or Class I, Zone 0 for US/Canada
  • Zone 0 for ATEX and IECEx
The evaluation and approval was performed by a third-party testing facility.
Is the Form Wash L ATEX certified or ATEX compliant?
The Form Wash L is not certified by a third party to meet ATEX standards. Electronic components located within the wash bucket, specifically the solvent monitor, are designed for compliance with ATEX standards.
How do typical Form Wash L IPA vapor levels compare to regulatory limits?
The safety limits for isopropyl alcohol (IPA) concentration in air as it pertains to health and flammability are provided in the table below. VOC levels may vary greatly during operation of the Form Wash L. Perform an on-site assessment to ensure that regulatory limits are not exceeded.
Isopropyl Alcohol (IPA) Vapor Concentration Levels
ValueExplanationConcentration
Lower Explosive LimitThe concentration which may ignite in the presence of an ignition source20,000 ppm1
IDLHThe concentration which is considered to be Immediately Dangerous to Life and Health2,000 ppm1
OSHA Permissible Exposure Limits (PEL)The concentration permissible for personnel exposure over an eight hour period400 ppm (over an eight hour period)1
Odor ThresholdThe concentration at which the odor of IPA can normally be detected40–200 ppm2
Footnote 1: US CDC National Institute for Occupational Safety and Health
Footnote 2: US National Center for Biotechnology Information
Below is a diagram of nominal IPA concentrations measured around the Form Wash L while a wet part is drying in the basket. Formlabs recommends ventilating your work area so that measured IPA concentrations do not exceed these guidelines.
alttext